Let's assume I have a bad calibration or that the app doesn't properly use the previous calibration.
When in calibrate page, the "uncalibrated ph reading" should show the ph ignoring any previous calibration, something like a raw value, right ?
if yes, than the app is ignoring the previously made calibration because the value shown in the calibration page is the same as the value shown on dashboard (that should be corrected based on the calibration)
I've checked in DB and both value for "cal1" and "cal2" are set and "calType" is set to 2
Posts: 3,251
Threads: 90
Likes Received: 163 in 120 posts
Likes Received: 163 in 120 posts
Reputation:
184
Joined: Sep 2015
Country:
Yeah the pH value on calibration page is a raw value so it should be different than home page if calibration is being applied. What do you have for buffer 1 and buffer 2 and cal1 and cal2? calType is correct for 2 point calibration. Maybe try deleting calibration, then verify in database buffer1 is 0 (equals 7.0ph) and cal2 is -177.48 (equals 10.0ph). Cal1 and cal2 should match at this point. After calibration cal1 and cal2 should be slightly different than the buffer values.
Posts: 164
Threads: 12
Likes Received: 14 in 14 posts
Likes Received: 14 in 14 posts
Reputation:
14
Joined: Oct 2021
Country:
02-16-2022, 01:16 AM
(This post was last modified: 02-16-2022, 01:16 AM by gandalf.)
this is what i have after a calibration but still same value shown on calibration screen and dashboard
(please add a 3 points calibration, with pH 4 buffer)
Posts: 3,251
Threads: 90
Likes Received: 163 in 120 posts
Likes Received: 163 in 120 posts
Reputation:
184
Joined: Sep 2015
Country:
That data looks good, I put it in mine and it adjusted pH as expected and when I went to calibrate it did change to the raw value. When you first load the calibration screen it will show the same as dashboard but the next update in 3 seconds should show the raw.
Try changing the calType from 2 to 0. 0 is no calibration, 1 is 1 point. After you change it you need to either restart the robotank.service or restart the Pi.
Posts: 164
Threads: 12
Likes Received: 14 in 14 posts
Likes Received: 14 in 14 posts
Reputation:
14
Joined: Oct 2021
Country:
Rob any update about the new version with improvements on dosing pumps?
i'm almost ready to put Robo Tank on, but i need a working dosing system, the current one is not really good as i cant slow down the pumps and it dose too much
Posts: 6
Threads: 2
Likes Received: 2 in 2 posts
Likes Received: 2 in 2 posts
Reputation:
2
Joined: Dec 2020
Country:
Is there an option to configure a push button for triggering a feed mode macro/script ?
Posts: 3,251
Threads: 90
Likes Received: 163 in 120 posts
Likes Received: 163 in 120 posts
Reputation:
184
Joined: Sep 2015
Country:
Sorry not sure when the dosing pumps will be finished, hopefully not much longer.
At this time there is no method to use a momentary push button, it needs to be a toggle but I will be adding that at some point.
Posts: 164
Threads: 12
Likes Received: 14 in 14 posts
Likes Received: 14 in 14 posts
Reputation:
14
Joined: Oct 2021
Country:
but what do you estimate? a couple of days? weeks? months?
Posts: 3,251
Threads: 90
Likes Received: 163 in 120 posts
Likes Received: 163 in 120 posts
Reputation:
184
Joined: Sep 2015
Country:
It'll be weeks till a month or two. It's very hard to estimate.
Posts: 164
Threads: 12
Likes Received: 14 in 14 posts
Likes Received: 14 in 14 posts
Reputation:
14
Joined: Oct 2021
Country:
02-21-2022, 01:28 AM
(This post was last modified: 02-21-2022, 01:29 AM by gandalf.)
ok so i'll wait. in one previous post you wrote "next week" so as this is the "next week" i was hoping for a new release
anyway, do you know how to make the pump go slower like reef-pi? i would like to make a patch, just to be able to dose 1ml or less. actually with robo-tank the pumps are running at full speed
Posts: 3,251
Threads: 90
Likes Received: 163 in 120 posts
Likes Received: 163 in 120 posts
Reputation:
184
Joined: Sep 2015
Country:
Yeah I was planning to have the v2/v3 compatibility done by now and then start on the dosing pumps. I was actually getting it ready to send out the other day and happened to test a DHT22 temp/humidity sensor which didn't work. I've been using the wiringPi library to manage the GPIO's and just removed that library to use sysfs instead but this broke the DHT22. The timing is critical when reading these, I suspect that's the problem. Spent many hours trying to get it working but can only with wiringPi. Very disappointing as I was so happy to remove that library.
The dosing pumps are set to run at full speed but the system is setup to be adjustable so it wouldn't be much to add a quick setting on the current page to tie over till I get the system changed. I'll shoot for next weekend as I don't want to put out an update with the DHT22 broke, I wish I didn't test it lol...
Posts: 164
Threads: 12
Likes Received: 14 in 14 posts
Likes Received: 14 in 14 posts
Reputation:
14
Joined: Oct 2021
Country:
Posts: 3,251
Threads: 90
Likes Received: 163 in 120 posts
Likes Received: 163 in 120 posts
Reputation:
184
Joined: Sep 2015
Country:
Hi Gandalf, sorry another week came and went. I have been putting some time into it and getting closer but need a little more time. I am hoping this week I can get it wrapped up, maybe it can happen by tonight.
Posts: 164
Threads: 12
Likes Received: 14 in 14 posts
Likes Received: 14 in 14 posts
Reputation:
14
Joined: Oct 2021
Country:
Posts: 164
Threads: 12
Likes Received: 14 in 14 posts
Likes Received: 14 in 14 posts
Reputation:
14
Joined: Oct 2021
Country:
I think the dosing pump calibration should be changed in something easier: insted of asking for ms/ml, jsut dose a defined amount of seconds, in example 10 secs.
Then ask how many milliliters where dosed in that time. much easier.
Posts: 3,251
Threads: 90
Likes Received: 163 in 120 posts
Likes Received: 163 in 120 posts
Reputation:
184
Joined: Sep 2015
Country:
I can see that being easier, as is you need to run it many times as you tune. This would be once and then verify.
Posts: 164
Threads: 12
Likes Received: 14 in 14 posts
Likes Received: 14 in 14 posts
Reputation:
14
Joined: Oct 2021
Country:
yes, maybe not once ti be accurate, but would be easier anyway
Posts: 71
Threads: 6
Likes Received: 1 in 1 posts
Likes Received: 1 in 1 posts
Reputation:
1
Joined: Sep 2017
Country:
Hi Rob, do you think it'll be possible to add a small lcd (maybe I2C) to show some basic tank data? Maybe water temperature, ecc...
Just to have a quick check of water status...
Posts: 3,251
Threads: 90
Likes Received: 163 in 120 posts
Likes Received: 163 in 120 posts
Reputation:
184
Joined: Sep 2015
Country:
Hi Simon, it's possible but it would be a while. It might be easier to use a really small HDMI screen, then the home page can be edited to fit.
Posts: 6
Threads: 2
Likes Received: 2 in 2 posts
Likes Received: 2 in 2 posts
Reputation:
2
Joined: Dec 2020
Country:
what would be the negatives in using a tablet with wifi and some web explorer instead of finding an really small HDMI screen ?
|