02-15-2022, 06:57 PM
Yeah the pH value on calibration page is a raw value so it should be different than home page if calibration is being applied. What do you have for buffer 1 and buffer 2 and cal1 and cal2? calType is correct for 2 point calibration. Maybe try deleting calibration, then verify in database buffer1 is 0 (equals 7.0ph) and cal2 is -177.48 (equals 10.0ph). Cal1 and cal2 should match at this point. After calibration cal1 and cal2 should be slightly different than the buffer values.